          Error 411 on GE MPH50

          Tags: Radiology - R/F

          (8)
          (2)

          Alberto Pepe

          Error 411 on GE MPH50

          November 01, 2020 11:05

          Hi,
          do you have information about the mph50 generator error 411, FPS_FAULT signal; status error?
          The voltage is symmetrical and correct, but error 411 appears first and then 513.

          1 person liked this
          Log inor Register

          to rate and post a comment

          (92)
          (33)
          (1)

          Konstantin Kuznetsov

          re: Error 411 on GE MPH50

          November 02, 2020 10:21

          MPH Code: 411
          Internal Code: FPS_ERROR
          Cause: Inverter floating power supply
          default (fps_fault signal)
          Effect: Exposure inhibited
          Action:Reset error.
          If it reappears, find the default in running diagnostics
          on/off and DC bus
          MENU2 TEST F1
          MENU2 TEST F3
          Then follow the faulty tree ON/OFF

          MPH Code: 513
          Internal Code: HE_INVERTER
          Cause: Inverter safety is set out of exposure (safety, kv_inf_kv0,
          exp_time_max, restarting_safety, kv_drop,
          kv_drop_an, kv_drop_cat, kv_max, mA_max, regul_out,
          i_inv_max, kv_sup_75kvn signals)
          Effect: Exposure inhibited
          Action: Reset error.
          If it reappears, find the default in running diagnostic kV
          MENU2 TEST F6
          MENU2 TEST F7

          The service manual is available in the Documentation Library / Customer Documentation Portal on the GE Healthcare official website. Document number 2165118-1EN. There is generally a lot of things available)

          Log inor Register

          to rate and post a comment

          (8)
          (2)

          Alberto Pepe

          re: re: Error 411 on GE MPH50

          November 06, 2020 11:01

          Hi Konstantin,
          I have the service manual and the schematics, but I don't have the service software to check.
          From the measurements made, there are no discrepancies on the tensions, they are stable and symmetrical.
          Besides that, I have problems on the ARC2000 starter, of which I am missing the schematics, can you help me?

          Log inor Register

          to rate and post a comment

          (1)

          eusebio rocha

          re: Error 411 on GE MPH50

          November 02, 2020 10:23

          i have no't the manual for this equipment, I sorry.

          Log inor Register

          to rate and post a comment

          Diagnostic Imaging Training: RSTI (Radiological Service Training Institute)

          15,000 service professionals from over 50 different countries have trained in over 35 courses in radiology, mammography, CT, ultrasound, networking, PACS, & DICOM. Classroom & lab facilities to accommodate up to 60 students.

          DOTmed Certified
          advertisement

           

          one source

           

          Hitachi

           

          RSTI Newsletter
          mg4155mg娱乐网址