Hi,
do you have information about the mph50 generator error 411, FPS_FAULT signal; status error?
The voltage is symmetrical and correct, but error 411 appears first and then 513.
MPH Code: 411
Internal Code: FPS_ERROR
Cause: Inverter floating power supply
default (fps_fault signal)
Effect: Exposure inhibited
Action:Reset error.
If it reappears, find the default in running diagnostics
on/off and DC bus
MENU2 TEST F1
MENU2 TEST F3
Then follow the faulty tree ON/OFF
MPH Code: 513
Internal Code: HE_INVERTER
Cause: Inverter safety is set out of exposure (safety, kv_inf_kv0,
exp_time_max, restarting_safety, kv_drop,
kv_drop_an, kv_drop_cat, kv_max, mA_max, regul_out,
i_inv_max, kv_sup_75kvn signals)
Effect: Exposure inhibited
Action: Reset error.
If it reappears, find the default in running diagnostic kV
MENU2 TEST F6
MENU2 TEST F7
The service manual is available in the Documentation Library / Customer Documentation Portal on the GE Healthcare official website. Document number 2165118-1EN. There is generally a lot of things available)
Hi Konstantin,
I have the service manual and the schematics, but I don't have the service software to check.
From the measurements made, there are no discrepancies on the tensions, they are stable and symmetrical.
Besides that, I have problems on the ARC2000 starter, of which I am missing the schematics, can you help me?
15,000 service professionals from over 50 different countries have trained in over 35 courses in radiology, mammography, CT, ultrasound, networking, PACS, & DICOM. Classroom & lab facilities to accommodate up to 60 students.
Alberto Pepe
Error 411 on GE MPH50
November 01, 2020 11:05
Hi,
do you have information about the mph50 generator error 411, FPS_FAULT signal; status error?
The voltage is symmetrical and correct, but error 411 appears first and then 513.
to rate and post a comment
Konstantin Kuznetsov
re: Error 411 on GE MPH50
November 02, 2020 10:21
MPH Code: 411
Internal Code: FPS_ERROR
Cause: Inverter floating power supply
default (fps_fault signal)
Effect: Exposure inhibited
Action:Reset error.
If it reappears, find the default in running diagnostics
on/off and DC bus
MENU2 TEST F1
MENU2 TEST F3
Then follow the faulty tree ON/OFF
MPH Code: 513
Internal Code: HE_INVERTER
Cause: Inverter safety is set out of exposure (safety, kv_inf_kv0,
exp_time_max, restarting_safety, kv_drop,
kv_drop_an, kv_drop_cat, kv_max, mA_max, regul_out,
i_inv_max, kv_sup_75kvn signals)
Effect: Exposure inhibited
Action: Reset error.
If it reappears, find the default in running diagnostic kV
MENU2 TEST F6
MENU2 TEST F7
The service manual is available in the Documentation Library / Customer Documentation Portal on the GE Healthcare official website. Document number 2165118-1EN. There is generally a lot of things available)
to rate and post a comment
Alberto Pepe
re: re: Error 411 on GE MPH50
November 06, 2020 11:01
Hi Konstantin,
I have the service manual and the schematics, but I don't have the service software to check.
From the measurements made, there are no discrepancies on the tensions, they are stable and symmetrical.
Besides that, I have problems on the ARC2000 starter, of which I am missing the schematics, can you help me?
to rate and post a comment
eusebio rocha
re: Error 411 on GE MPH50
November 02, 2020 10:23
i have no't the manual for this equipment, I sorry.
to rate and post a comment