          Siemens polyphos 30 m

          Tags: Radiology - R/F

          (1)

          Abdulbaset Alwashil

          Siemens polyphos 30 m

          October 29, 2020 10:20

          I have error 2 and 3 in seimens polyphos 30 m.
          Can any one help me ?

          Log inor Register

          to rate and post a comment

          (121)
          (7)
          (16)

          Imad Muati

          re: Siemens polyphos 30 m

          November 02, 2020 10:21

          Error 2 : no tube heating check Fuse D10 ( D18 ) , f2 D 13, D12-mp- Vhist (fil): 80 mv
          Error 3 : no basic heating check D13 B22 0 volt ?

          Log inor Register

          to rate and post a comment